help
help > RE: Window Width and Window Center Value of DICOM images not considered while converting dcm to nii
Sep 11, 2019 02:09 PM | Chris Rorden
RE: Window Width and Window Center Value of DICOM images not considered while converting dcm to nii
Your image has far more dynamic range than the screen. Set the
image intensity minimum and maximum to highlight the tissues you
are interested in. CT scans are saved as Hounsfield units, with the
intensity of air very dark (-1000), soft-tissue and water in the
range 0-50 and bone very bright (+1000). On option is to just set
the "min" and "max" values for the window values you desire.
Alternatively, right-drag your mouse over the area of the image you
are interested in and it will adjust the contrast to what you are
interested in.
If you are using [MRIcroGL](https://www.nitrc.org/plugins/mwiki/inde...) you can select one of the "CT" color schemes. The image below shows the CT_Abdo scan that comes with MRIcroGL (File/OpenStandard/CTAbdo) with both the initial intensity range and with the "CT_Kidneys" color scheme applied. Choose Scripting/Templates/ct_abdomen for a graphical demonstration.
If you are using [MRIcroGL](https://www.nitrc.org/plugins/mwiki/inde...) you can select one of the "CT" color schemes. The image below shows the CT_Abdo scan that comes with MRIcroGL (File/OpenStandard/CTAbdo) with both the initial intensity range and with the "CT_Kidneys" color scheme applied. Choose Scripting/Templates/ct_abdomen for a graphical demonstration.